LCR Meter Problem with Second Input offset adjustment
Hi!
I just build my LCR Meter and tried to do the first time calibration.
The first Input Offset adjustment worked without any problems.
But on the Second Input offset adjustment I always get measurements between 1.5V and 2V (the target is 0.00V).
Do you have any suggestion where the problem might be?
On TP5 I always measure 1.2mV during the adjustment.
On TP10 I have 106mV.
Difference between TP9 and TP11 after the failed adjustment is 1.35V (the AU2011 Program shows 1.535V).
Thanks
Best Regards
Benjamin