SiamCatI received the Elektor tested mainboard of the LCR-meter, put it in the alu-case, did the first-use procedure, all is well. Trim without Kelvin-clips goes well: closed with jumper (0,00 Ohm) and open (>1Gohm and 0,00pF).
But when I use the Kelvin-clips the open trim fails. Also the PC-display hangs, or changes far too often (tx/rx-leds do not flash or flash continuously), and values showed are strange: a few k-ohms to a few megs, L-values varying from 100H down to few micro-H's etc. It seems very unstable with Kelvin-clips. The closed trim with Klevin-clips goes well.
Is this issue known by other users? And what could be the cause of this problem?
The problem is with the power line.
Kelvin clips cables are, on the frequency of the electrical line, like antennas.
To reduce this phenomenon, I recommend:
- shorten the cables up (they are very long original).
- make sure that the housing is connected to Earth (it is not automatic with a laptop!)
- If you have the possibility, place a metallic sheet steel connected with the earth below the device and the cables.
The new versions of Firmware (version 3.0.0) and of the program AU2011 (version 3.0.0), allow to limit the maximum gain in range 8 (and also in range 1 to solve the problem of the measure of resistances < 1 ohm).
They should be made available on the site of Elektor very quickly.
Jean-Jacques